    Default Another Would You Bother Thread

    This time, hopefully nothing inciteful.

    This time Japanese versus Swiss, both used, both sometimes available relatively inexpensively. The former more so than the latter.

    Last weekend I had been fooling around with my .0001" Mitutoyo and experienced a minor variation of reading when running the stylus up a couple of thou onto a gauge block. As a result of this observation, Michael G very generously posted to me his Interapid and his Brown and Sharpe Bestest (TESA) dial test indicators for my humble appraisal. Both tenth indicators, used and in excellent nick.

    Surprisingly both Swiss suffered the same tiny variation in repeatability as the Mitutoyo did. Maybe .00002" - .00003". This is just my uneducated observation, others may have an explanation for the occurrence.

    Comparison of the physical attributes of the indicators is interesting. The Interapid ( reputedly the best DTI money can buy ) shares the same screw on dovetails as Mitutoyo use on their tenth DTI . My half thou Mitutoyos have integral dovetails machined into the indicator's body casting as does the Bestest. Compac, in their catalogue, make a point of informing the reader that their indicators all have integral dovetails. Different strokes.

    The winning feature in my mind is the hinged rear mounting on the Interapid. This allows the indicator to be both swiveled and very conveniently, rotated 360 degrees. A flexibility not found with any other DTIs I've handled.

    Interapid also offer an articulated mounting arm facilitating the most convoluted set ups. In defence of Mitutoyo, their universal DTI with its rotatable nose is a handy bit of kit, Stu and RC might agree.

    So, would I bother going Swiss? If I hadn't bought a 30 year old .002mm Mitutoyo test indicator two weeks ago, probably and probably an Interapid.

    BT
    Attached Images Attached Images

  2. # ADS
    Google Adsense Advertisement
    Join Date
    Always
    Location
    Advertising world
    Age
    2010
    Posts
    Many





     
  3. #2
    Join Date
    Oct 2011
    Location
    Adelaide
    Age
    59
    Posts
    3,149

    Default

    As I said to Bob at the time "Would I dare to send anything but Swiss to him to test?".

    Interesting that other tenth DTIs share the same characteristic. Thinking about the size of the numbers I'm now wondering whether it might be a mechanical play issue. A perfect fit should mean no play but there have to be mechanical clearances and tolerances so that may be how they are manifest.

    (And now that you have extolled their virtues I might have to use them more often and take better care of them)

    Michael

  4. #3
    Join Date
    Oct 2012
    Location
    Australia
    Posts
    765

    Default

    Quote Originally Posted by Anorak Bob View Post
    This time, hopefully nothing inciteful.

    This time Japanese versus Swiss, both used, both sometimes available relatively inexpensively. The former more so than the latter.

    Last weekend I had been fooling around with my .0001" Mitutoyo and experienced a minor variation of reading when running the stylus up a couple of thou onto a gauge block. As a result of this observation, Michael G very generously posted to me his Interapid and his Brown and Sharpe Bestest (TESA) dial test indicators for my humble appraisal. Both tenth indicators, used and in excellent nick.

    Surprisingly both Swiss suffered the same tiny variation in repeatability as the Mitutoyo did. Maybe .00002" - .00003". This is just my uneducated observation, others may have an explanation for the occurrence.

    Comparison of the physical attributes of the indicators is interesting. The Interapid ( reputedly the best DTI money can buy ) shares the same screw on dovetails as Mitutoyo use on their tenth DTI . My half thou Mitutoyos have integral dovetails machined into the indicator's body casting as does the Bestest. Compac, in their catalogue, make a point of informing the reader that their indicators all have integral dovetails. Different strokes.

    The winning feature in my mind is the hinged rear mounting on the Interapid. This allows the indicator to be both swiveled and very conveniently, rotated 360 degrees. A flexibility not found with any other DTIs I've handled.

    Interapid also offer an articulated mounting arm facilitating the most convoluted set ups. In defence of Mitutoyo, their universal DTI with its rotatable nose is a handy bit of kit, Stu and RC might agree.

    So, would I bother going Swiss? If I hadn't bought a 30 year old .002mm Mitutoyo test indicator two weeks ago, probably and probably an Interapid.

    BT

    What are the spec on repeatability meant to be? In my experience repeatability, hysteresis and measuring force is a real tricky thing to get right especially at the finer resolution eg 50 millionths and 0.5 um, everything gets in your way to making good measurements with low variability, body heat being the most persistent annoyance. I have only seen (never used) digital probes with good level of repeatability eg ± < 1 resolution division (eg Mahr Federal Millimar 1318).

    Hysteresis is also an important one to point out as this is "normally" a couple of divisions.

    -Josh

  5. #4
    Join Date
    Oct 2012
    Location
    Australia
    Posts
    765

    Default

    Oh the other thing that I seem to recall is that the intrepid is only true at 12 degrees?

  6. #5
    Join Date
    Apr 2012
    Location
    Melbourne
    Age
    57
    Posts
    122

    Default

    "Maybe .00002" - .00003". This is just my uneducated observation, others may have an explanation for the occurrence. "
    Your worried you need this sort of accuracy for what reason.
    Unless your using the equipment in a temperature controlled environment you won't be able to control variations like this anyway.

  7. #6
    Join Date
    Jun 2008
    Location
    Victoria, Australia
    Age
    74
    Posts
    6,132

    Default

    Quote Originally Posted by Clubman7 View Post
    "Maybe .00002" - .00003". This is just my uneducated observation, others may have an explanation for the occurrence. "
    Your worried you need this sort of accuracy for what reason.
    Unless your using the equipment in a temperature controlled environment you won't be able to control variations like this anyway.

    Yes absolutely correct.. just to put it in perspective, steel expands/contracts at 13 microns per meter per degree C, so for a 50mm part a one degree C change will change the size of the part by 0.65 microns, ( 0.050 * 13 )

    So to see a 0.00002" change ( that's 20 microinches, which is 0.5 microns ) would only require less than 1 degree C change in temperature of the part.

    Keeping heat sources like hands away and shielding from draughts and air currents, is not always easy..

    Regards
    Ray

  8. #7
    Join Date
    Nov 2008
    Location
    Perth WA
    Age
    71
    Posts
    5,650

    Default

    Quote Originally Posted by Clubman7 View Post
    "Maybe .00002" - .00003". This is just my uneducated observation, others may have an explanation for the occurrence. "
    Your worried you need this sort of accuracy for what reason.
    Unless your using the equipment in a temperature controlled environment you won't be able to control variations like this anyway.
    It was an observation. I did not state that I had need for accuracy of that sort. You might, I certainly don't.

  9. #8
    Join Date
    Oct 2011
    Location
    Adelaide
    Age
    59
    Posts
    3,149

    Default

    Bob, have you slipped an extra zero in there? I've just realised that we are talking hundreds of thous.

    Michael

  10. #9
    Join Date
    Aug 2008
    Location
    near Rockhampton
    Posts
    4,304

    Default

    I am just not a big fan of old Mitutoyo gear... The modern stuff is good... The older stuff I think less good.. or probably more to do with just the bad luck I have had with older mit indicators...
    Light red, the colour of choice for the discerning man.

  11. #10
    Join Date
    Apr 2012
    Location
    Melbourne
    Age
    57
    Posts
    122

    Default

    Quote Originally Posted by Anorak Bob View Post
    It was an observation. I did not state that I had need for accuracy of that sort. You might, I certainly don't.
    "This is just my uneducated observation, others may have an explanation for the occurrence. "

    You asked for an explanation.

  12. #11
    Join Date
    Nov 2008
    Location
    Perth WA
    Age
    71
    Posts
    5,650

    Default

    Quote Originally Posted by Michael G View Post
    Bob, have you slipped an extra zero in there? I've just realised that we are talking hundreds of thous.

    Michael
    OK, I was looking at the dials with a 12X loupe screwed up against my eyeball. At that magnification the space between a one tenth graduation is pretty large, large enough to have a stab at the needle position varying .00002'' to .00003'' and I wondered why the variation was occurring. In the case of my Japanese indicator, wear would probably a influencing factor, possibly the same with your Swiss indicators.

    I have no illusions about the irrelevance of measurement at this level to me. My shed shares more in common with a tent than it ever would a toolroom.

    Mention a micron and the ears prick up. It's a bit like "who does he think he is using the talk of the toolroom."

  13. #12
    Join Date
    Apr 2012
    Location
    Melbourne
    Age
    57
    Posts
    122

    Default

    Bob, your talking 0.0005 to 0.0007 of a mm variation.
    You might find the manufacturing tolerances of the indicators you tested wouldn't
    be this fine when manufactured.

  14. #13
    Join Date
    Oct 2012
    Location
    Australia
    Posts
    765

    Default

    Quote Originally Posted by Anorak Bob View Post
    OK, I was looking at the dials with a 12X loupe screwed up against my eyeball. At that magnification the space between a one tenth graduation is pretty large, large enough to have a stab at the needle position varying .00002'' to .00003'' and I wondered why the variation was occurring. In the case of my Japanese indicator, wear would probably a influencing factor, possibly the same with your Swiss indicators.

    I have no illusions about the irrelevance of measurement at this level to me. My shed shares more in common with a tent than it ever would a toolroom.

    Mention a micron and the ears prick up. It's a bit like "who does he think he is using the talk of the toolroom."
    You may also be looking at surface deformation of the gauge block and the tip. This always happens to me using a 0.0005mm probe on steel/CI, but also is probably because it has quite a high measuring force; I think it is about 0.1-0.05N.

    Also I note in the photo you still have some trace of blue on the rock.

    Try gloves, breath/body shield, "super clean"/pristine gauge blocks, "super clean" surface plate, and repeat. To be honest, I'm impressed they hold up as well as they do, as I have said hysteresis is normally at least a division if not 2.

Similar Threads

  1. Should I bother with a bent board?
    By GarciaJ in forum WOODWORK - GENERAL
    Replies: 17
    Last Post: 3rd April 2013, 11:28 PM
  2. where to get a good pom-pom gun (and should I bother)
    By Michael G in forum METALWORK FORUM
    Replies: 13
    Last Post: 24th July 2012, 10:16 AM
  3. Don't bother with Disston saws
    By MajorPanic in forum HAND TOOLS - UNPOWERED
    Replies: 11
    Last Post: 25th September 2011, 10:56 PM
  4. Do restorers bother with tinted wax ?
    By JDarvall in forum FINISHING
    Replies: 18
    Last Post: 22nd August 2009, 03:53 PM
  5. oh bother
    By mic-d in forum TRITON / GMC
    Replies: 0
    Last Post: 16th September 2004, 06:02 PM

Tags for this Thread

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •