    Default Air flow calibration measurements

    This is a technical post about measuring air speed/flow, which I know won't interest many woodies but it does highlight the many difficulties involved in making reliable air flow measurements.
    It involves a cross comparison of 3 different air flow meters. This is just part one of a multipart post which will evolve in this thread over the next few weeks

    I recently purchased a Testo hot wire anemometer (air speed meter) and was interested to see how well it agrees with my 2 other hot wire anemometers.
    Lappa has posted in this forum on using a Testo to measure airflows in his DC system.

    The Testo is a very handy product since it can be left operating in a remote location in a ducting system while recording of data takes place on a tablet or mobile phone communicating with the probe via Blue tooth.
    It comes with a one point calibration certificate which states at 8.0 m/s it was measuring 8.2 m/s and has an absolute uncertainty of +/- 8.8%.
    Knowledge of this absolute uncertainty is critical in comparing readings obtained by different sensors.

    The two other meters I am comparing the Testo to are

    1) Kurz meter.
    This is an old analog meter with a zero to 30 m/s range and comes with a calibration certificate stating it has been compared to a NIST standard and has an uncertainty of +/-4%, but given how old it is I really doubt it's that good.
    It jumps around quite a bit while measurements are made and you will see how I got around this problem below.

    2) TSI meter
    This is a much newer digital meter with a zero to 20 m/s range and comes with a 10 point calibration certificate traceable to NIST standards. It does have a short term integration feature which smooths the readings. The quoted uncertainty is +/- 5% but the calibration curve shows that the accuracy is > +/-2.5% for all readings above 60 FPM. Because of the quality of this calibration I decide to reference the Testo and the Kurz against the TSI meter.

    Simple comparative tests have been done before by me between the TSI and the Kurz many years ago when I convinced myself they agreed within spec.

    A real world test would be to use each of the meters to perform a full air flow (CFM) test in the same duct at the same air speed. Each measurement involves the systematic measurement of the air speed (m/s of FPM) at different radial points across duct and then performing calculations to get the flow (CFM). I will eventually do this but for the moment I wanted to see if I could get a comparison of just the air speed using another method.

    To do this I needed to setup a test duct and locate the sensor in the same relative point in the air stream so the 3 probes measure the same air speed. This is not as easy as it sounds as the air inside a normal duct varies speed with distance from the walls of the duct and also (especially at higher speeds) preferentially weaves and wanders around in the the duct. If the sensors are placed serially along the duct they cannot be too close together as they will interfere with each other. If serial placement was used the would need to be placed meters apart and given the test duct for one sensor needs to be around 10x as long as the duct is wide, this would require a very long test duct.

    The other arrangement is with all 3 sensors at the same position along a duct but distributed evenly (120º apart) at the same radius so the sensor elements are all the same distance from the wall of the duct and this is what I did. .

    Probes.jpg

    A complication arises in that the Testo probe has a 9.1mm diameter while the TSI and Kurz are 1/4" diameter probes.
    This alone could change things significantly so I turned up some 9.1 mm diameter Al sleeves that were placed over the TSI and Kurz sensor shafts but I had to leave the ends of the sensor uncovered others I would disturb the calibration of the TSI and Kurz. This is major unknown effect #1.

    The duct itself consists of two lengths of 9" (240mm) diameter PVC stormwater pipe.
    The reason I used two pieces was because the sensors had to be located in the middle of the length of the duct and I needed to be able to reach in to make sure they were the same distance away from the wall of the duct. This I was able to do with a tolerance of <1%. I also had to be able to see if the sensor heads were oriented correctly. Being able to break the duct and get up close to the sensors was the only way this was easy to do.

    Despite the great pains taken to get the probes at the same distance from the walls of the duct (113 mm) this is no guarantee they will experience the same air speed - see below,

    Here is what the test duct looks like
    It's around 4m long with the probes placed at the 2m mark.
    For accuracy the test duct length should be ~10x the duct diameter. Even better would to have a length 10x the duct diameter before the sensors and another length 10X the duct diameter after the sensors.
    This would require ~2400 mm before the seniors and 2400 mm after the sensors. My setup has 2000 mm before and 2000 mm after the sensors, so technically I am a little short but I am deliberately working at slow air speeds for this preliminary test so it should be OK, although it remains as a minor unknown. A good test is how the air speed varies in time which I will discuss later.

    Testduct2.jpg

    The piece of duct with the probes in it can be disconnected and rotated to see if there is a positional effect. This is necessary to check because the flow can differ inside the duct due to back reflections from junctions etc down the line. This involves lengthy systematic testing that I will do next.

    Another view showing the pot plant reducer that attaches to the ducting for my lathe
    In the middle you can see an MDF ring which is where the sensors attach to from the outside.
    Testduct.jpg


    Here is a close up of the MDF ring with the Testo probe (Orange and black) on top and the Kurz probe at the 4 o'clock position.
    The TSI is at the 8 o'clock position on the other side of the duct.
    The small screw in the middle of the Al strip straddling the probe, locks the probe in place.
    sensorplacement.jpg

    Now you have 3 meters jumping around like 3 frogs in a sock so how the hell do you record each reading at exactly the same time.
    Ideally you plug all 3 into a computer and even the old Kurz has an analog signal out cable that was used to connect it to a computer but I have lost the calibration data for this and I did not want to go through that calibration again so I had to find another way.
    I've done this a few times before and all I do is get a digital camera and take a photo of all 3 of the displays.
    I just click the camera 10-15 times at the same air speed, open or close the gate valve to the lathe to vary the air speed and repeat the measurements.
    The air speeds I used did not matter as long as a range of speeds was used during the test.

    Then I review these on the computer and type them into a spreadsheet - it sounds long winded but its actually quite fast and it is a good chance to get up close and personal with the data rather than blindly using what comes out of the computer.
    It's also a chance to reject wild outliers and start to make sense of the data

    Testductprobes.jpg

    I took 10 sets of readings between 1 and 9 m/s (with 2 sets at ~6 m/s)
    The 10-15 data points for each set are averaged and age results for each meter compared to each other.
    The "error bars" are a measure of the overall tolerance/uncertainty in the comparison for each point.
    These are the simple sum of the standard statistical error for the TSI meter measurement and the standard statistical error for the other meter .
    These should really be added in quadrature (square root of the sum of the squares but I was being conservative in my uncertainty assessment.

    The air speed variation in time differs significantly between the 3 meters.
    The Kurz reports instantaneous speed with about a 1/4 second response rate so that had the greatest variability, about 50% more than the variability of the Testo.
    The TSI reported the least variability between readings, (about half that of the Testo) but that could be because the TSI does have a weak smoothing function built in.
    This should not affect the final results.

    The results are plotted below as % differences of the Testo and Kurz from the TSI readings.


    Screen Shot 2017-04-22 at 10.37.49 PM.png

    What does it all mean?
    A simple conclusion is, if the error bars for a point on the graph cross the zero vertical axis the meter agrees within spec with the TSI meter for that air speed.
    For the Testo it appears to consistently read on the low side compared to the TSI meter, with all bar one of the readings below that of the TSI.
    However, 6 out of 10 of the Testo readings (1, 2.7, 3, 5 and two 6 m/s) are within uncertainty which shows they are not far from agreement.
    The average deviation for the Testo from the TDI is -8.5% (the "minus" means its reading too low) and given the absolute uncertainty of the Testo is +/- 8.8%, from that point of view it could be said the two meters agree within spec.

    The Kurz has an average deviation of -6.5% which while slightly closer than the Testo to the TSI is outside the Kurz quoted spec of 4%, but given how old this meter is, this does not surprise me.

    Overall, given the gotchas, I am amazed these results got anywhere near agreeing to this degree, and there is much more testing to do especially at higher speeds.
    The kind of test described in this post cannot be done at higher air speed because to achieve this will require the use of 6" duct (to get the air speeds up) and the smaller cut will put the probes closer together which is likely to create a mess of the air flow.
    The only way to do sensor comparisons at these speeds is to perform full airflow measurements.

    One possible outcome of this experiment is that if the Testo is indeed reading on the low side it could be that Lappa's system is better than he thinks.
    And who's to say my Testo is the same as Lappa's in the first place?
    Conversely the TSI could also be reading too high and I could have a slightly worse system than I think I do.

    Now let me reiterate that the above conclusions could still be in doubt because I have not really assessed the chances of non-radial variation of air speed within the duct. This could easily result in the deviations observed.
    To test this I will be rotating the section of duct that holds the sensors into a different positions with different sensors at the top each time and repeating these measurements.
    Even this might not show up everything but it will be a start.
    The real test will be a full volume measurement for which I will have to make up shorter Al sleeves for the TSI and Kurz meters.

    What it does show is how different meters can give different results and how messy in general air flow measurements are.
    If this doesn't put you off air flow measurements I think by the end of this thread you may convinced.

  2. # ADS
    Google Adsense Advertisement
    Join Date
    Always
    Location
    Advertising world
    Posts
    Many





     
  3. #2
    Join Date
    Aug 2007
    Location
    Saskatoon, SK, Canada.
    Posts
    1,439

    Default

    I'm looking forward to seeing what the results are overall when you are done. Given any thought to see how much of a difference to the reedings when you put a bell mouth on the duct?

    Pete

  4. #3
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    Quote Originally Posted by QC Inspector View Post
    I'm looking forward to seeing what the results are overall when you are done. Given any thought to see how much of a difference to the reedings when you put a bell mouth on the duct?

    Pete
    Ah ha - I have been waiting to do that for a while.
    I did some tests with the first MDF turned BMH I made a few years ago with a jerry rigged testing setup and the results showed a small modest increase, but while I have this testing setup in place it's one thing I think I can test a bit more reliably than last time. This test only requires a relative comparison, i.e. with and without the BMH, so absolute accuracy is not required.

  5. #4
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    OK I have rotated the test duct , so the 3 sensors alternately sit in the 12, 4 and 8 o'clock positions, and repeated the calibration.
    This is being done to see if there are preferential airflow paths in the test duct.

    I'll show only the Testo calibrations against the TSI meter otherwise the graphs gets very crowded.
    The blue Testo12 line and points are for the Test at the 12 o'clock position , Testo4 is when the Testo is at the 4 O'clock position etc.

    The XTesto12 point is the average Deviation across the range for all the 12O'clock deviations, etc.

    Screen Shot 2017-04-23 at 9.39.29 PM.png

    Although the data is somewhat scattered, the results for the 4 and 8 O'clock positions appear a little closer to zero deviation line than the 12 O'clock position, but otherwise the averages appear to overlap so much that one cannot really tell the difference.

    My summary is-
    There are no discernible preferential air flow paths in the duct
    My Testo, between air speeds of 1 and 9 m/s, measures on average about 5% lower than the TSI meter.
    OR
    The TSO measures 5% higher than than Testo meter.

  6. #5
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    I did a quick and dirty BMH measurement - not a full airflow (CFM) measurement - just what the difference in air speed is

    without (just a short stub of 150mm duct)

    IMG_7744test2.jpg

    and with the BMH in place
    IMG_7744test.jpg

    20 readings on each were averaged.

    The TSI meter gave a 15.8% increase while the Testo gave a 11.7% increase
    The uncertainties were +/- 1.0 and +/- 2.1 % respectively.
    Note the result means they don't agree but it was a quick and dirty measurement only of the air speed in one place in the duct.
    A full air flow (CFM) measurement would be more definitive.

    None the less it's clear there is a difference - it's the easiest and cheapest 10% improvement in flow you'll get.

    The change in motor current is only 0.1A and as the built in ammeter on the DC remote switch box only reads to 0.1A its rather surprising.

    I don't know how much longer I can keep the test ducting in place in my shed as it takes up a lot of room and divides the shed in half.
    It's going to be used at the mens shed in the near future.

  7. #6
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    I did an even quicker and dirtier change in air speed with/without an oversized 4" BMH.

    Now things are really hanging out in the breeze 1) 9" duct, 2) 9 - 6" reducer, 3) 6" duct, 4) 6 - 4" reducer, and then 5) 4" BMH
    Its all just "push fit" but things are starting to get "wobbly".

    And . . . . . .

    I noticed that even gently adding and removing the BMHs to the end of the other bits and pieces on the end if the test pipe changes the air speed/direction by a visible amount.
    This highlights the problems of measuring air speed and the need to measure total air flow.

    To make sure I was not changing the air speed/direction, I went back and forth (BMH/no BMH/BMH) a few times to check I had not changed things
    The result was 17% improvement!!! - the flow looks pretty good too.

    Tomorrow I am going to clamp and screw the bits and piece on the end of the flange down to make sure I'm getting reproducible results.

  8. #7
    Join Date
    Jan 2009
    Location
    Brisbane
    Posts
    1,770

    Default

    Thank you for doing this work.

    Cheers
    There ain't no devil, it's just god when he's drunk!!

    Tom Waits

  9. #8
    Join Date
    Aug 2007
    Location
    Saskatoon, SK, Canada.
    Posts
    1,439

    Default

    Interesting to see the difference a smooth intake makes. Actually when I suggested a bell mouth, I was thinking a 9" one. I thought it would reduce turbulence of the air at the probes?

    Pete

  10. #9
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    Quote Originally Posted by QC Inspector View Post
    Interesting to see the difference a smooth intake makes. Actually when I suggested a bell mouth, I was thinking a 9" one. I thought it would reduce turbulence of the air at the probes?

    Pete
    Yes it should.
    Making one is not going to be easy.
    A 250 mm BMH will need to be 375 mm in diameter - I will need a bigger lathe to make up the former
    Anyone have a lathe with a 16" swing I could borrow for a couple of hours. I can be an outboard swing because a tail stock is needed to drive the 250mm pipe onto the former

  11. #10
    Join Date
    Aug 2007
    Location
    Saskatoon, SK, Canada.
    Posts
    1,439

    Default

    Time for a bigger flower pot.

    Pete

  12. #11
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    I spent some of today rejigging the test pipe so I could make volume flow (up until now I have just been doing air speed) measurements using any of my meters.
    I used small tek screws and pipe clamps to lock everything in place - even so one has to be careful not to knock the set-up otherwise the results can change quite dramatically.

    The first test I did was on the 6" BMH with the Testo meter and I got an increase in flow of 10.4 to 10.8% which agrees with the quick and dirty "air speed only measurement".

    To minimise the number of changes and to get more comparable results I'd set the probe to a prescribed radial position and then measure the air speed "with" and then "without" the hood; then move the sensor position and then measure without and with the hood. This reduces the number of times the hood is removed and replaced by 50%. I did this all the way across the test duct.

    The Testo is real nice to use - I can use the meter in my shed via my iPad mini, and after each measurement email the data to my PC up in the house.
    Even so the full comparative volume/flow test takes about 45 minutes to collect the data, and then it takes about a hour to process all the data.

    It was interesting to see the air speed difference between the "BMH" and "no BMH" consistently between 8 about and 10% across the middle of the pipe.

    Screen Shot 2017-04-25 at 8.46.09 PM.png

    At about 1/4 of the way across the pipe (3cm mark) the air speed difference increased to about 20% and then dropped to near zero when the meter probe was close to the internal wall of the duct where the air speed is quite low anyway.

    Because of the air speed differences being so consistent in the middle of this test duct I think I can test things out in terms of % difference just using a couple of air speed measurements. This will make testing a lot quicker than doing a complete volume/flow measurements. if things appear too close to distinguish I may still need to use volume/flow measurements.

  13. #12
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    This norming I did a couple of interesting tesst.

    I did not realise that when I did the above test yesterday it was being done with a second 6" port wide open at the same time, so the volume/flow in the test duct was around 500 CFM.

    This morning I checked it again with only the one 6" port open and got a 12.5% improvement with the BMH.

    I also measured the effect of the BMH at around 1/4 flow and got an improvement of 3.1 + 1.7%

    This makes sense as the greater the flow the more effective you would expect the BMH to be. Or maybe this only works up to a certain point?

    Also I tested the effect of the 1" mesh guard at full flow and the resultant loss in flow is 1.14 +/- 0.84% , so just discernible, but more importantly it does not substantially negate the effect of the BMH.


    Bell mouth hood offer - expressions of interest-hanger-jpg

    Now I'm going to try some 100 mm BMHs.

  14. #13
    Join Date
    Nov 2013
    Location
    Caboolture QLD AU
    Posts
    781

    Default

    I'm wondering if some users reading the above may think that 10% to 12% is not a great enough improvement to justify making buying or fitting a bell mouth. However what I find with the Bell mouth inlets (as Bob has mentioned before) is a greatly enhanced (larger) area of pickup forward of the bell mouth, way more dust capture area, now add that 20% increase in flow and the use of a bell mouth intake is, IMHO, absolutely worth the effort. I've converted 3 x 150mm intakes with the Bell mouths I got from Bob, and have just fitted the 100mm unit to the above table pickup on the Bandsaw, once again, a noticeable improvement, even on the tiny bits of white laminate that occasionally pop off.

    Look forward to seeing how the 100mm BM shapes up in the tests.

    Forgot to add: Really great to have the small loss with the intake screen verified at last.

  15. #14
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    Quote Originally Posted by MandJ View Post
    I'm wondering if some users reading the above may think that 10% to 12% is not a great enough improvement to justify making buying or fitting a bell mouth. However what I find with the Bell mouth inlets (as Bob has mentioned before) is a greatly enhanced (larger) area of pickup forward of the bell mouth, way more dust capture area, now add that 20% increase in flow and the use of a bell mouth intake is, IMHO, absolutely worth the effort. I've converted 3 x 150mm intakes with the Bell mouths I got from Bob, and have just fitted the 100mm unit to the above table pickup on the Bandsaw, once again, a noticeable improvement, even on the tiny bits of white laminate that occasionally pop off.

    Look forward to seeing how the 100mm BM shapes up in the tests..
    100mm BM measurements have been done - just taking a break from processing the data results coming soon. You are right about the advantage being much more than just the actual flow increase. It captures much more dust from the front of the hood so if you point it at a dust source it picks up more of that dust and not just air with little or no dust in it from the sides and behind the hood, which is what a naked duct doe.

  16. #15
    Join Date
    Feb 2006
    Location
    Perth
    Posts
    27,796

    Default

    Some more tests.

    This (the yellow one) was the first 150 mm BMH I made from 3 x 18 mm thick layers of MDF. It was not a neat fit on the elbow and I only every did a basic test on its performance and I can't even remember why I go for it - somewhere between 5 and 10%

    The improvement if flow for this one at full flow is 8 +/- 1%
    It has been replaced by a PVC type.

    Improving machine cabinet dust ports-img_1878-jpg

    This was the first of the 100 mm MDF BMHs I made.
    Its a bit more trumpet shaped than the PVC ones and gave a 20.4 +/- 1.4 % improvement in flow


    Ducting update.-bmhguard-jpg


    This standard PVC 100 mm BMH gave an 18.3 +/- 1.2% improvement.
    More bandsaw dust extraction ideas.-img_1919-jpg
    These of course are flow dependent. If you use it on a 1HP DC with clogged bags you won't see that much of an improvement

    Now - what else can I test?

Page 1 of 4 1234 LastLast

Similar Threads

  1. calibration of a cnc
    By mat_au in forum CNC Machines
    Replies: 7
    Last Post: 30th September 2015, 12:13 PM
  2. Some interesting air flow measurements
    By BobL in forum DUST EXTRACTION
    Replies: 4
    Last Post: 18th May 2013, 12:33 PM
  3. Calibration, settings and tuning
    By TritonJapan in forum TRITON / GMC
    Replies: 3
    Last Post: 15th November 2006, 05:12 PM
  4. Dusty air flow - some measurements
    By BobL in forum SAFETY
    Replies: 14
    Last Post: 25th May 2006, 02:02 AM
  5. Moisture meter calibration
    By necy in forum WOODWORK - GENERAL
    Replies: 4
    Last Post: 21st May 2006, 11:06 PM

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •